Is there any molecular vibrational mode (or activity) that is neither Raman nor IR active?

In other words. Is there any activity in the molecules that can't be captured by either Raman spectrometer (all wavelenght) or IR absorption spectrometer (all wavelenght)? What are those?

Jtl
  • 577
  • 4
  • 9
  • 2
    Why, there are many such modes (silent, as they are called). – Ivan Neretin Oct 07 '19 at 11:14
  • Please give some examples of what they could be. Thanks. – Jtl Oct 07 '19 at 11:23
  • 2
    A molecule's normal modes are determined by its symmetry, the selection rules for transitions by change in dipole (IR) or change in polarisability (raman) which can only be satisfied by certain normal modes. If you look at point groups tables there are many irreps (symmetry species) with neither ($x,y,z$) for IR, ,or product operators ($xy, x^2$, etc), Raman. The silent vibrational frequency can sometimes be worked out by comparing the measured heat capacity with that produced via the partition function using the measured frequencies, the discrepancy is the unknown vibrational energy. – porphyrin Oct 07 '19 at 21:11

1 Answers1

8

Yes it is possible. While no normal modes can be both IR and Raman active in molecules with a centre of symmetry, as mentioned in a comment molecules with silent modes exist.

The easiest example is probably the HCH out-of-plane twisting of ethylene: enter image description here

Source: http://www.shodor.org/succeed-1.0/compchem/labs/vibrations/ethylene.jpg

Due to the symmetry of the molecule, when the HCH twist out of plane the dipole and polarizability of molecule is unchanged.

Other examples can be found here

https://www.chem.purdue.edu/jmol/vibs/c6h6.html

where the modes of benzene are sketched. The silent modes easily seen are kind of distorted breathing of the ring, that due to the C6 axis of symmetry again let the molecule unchanged both for the electric dipole moment and its polarisability.

Karsten
  • 40,170
  • 8
  • 67
  • 183
Alchimista
  • 3,626
  • 2
  • 9
  • 16
  • Does it mean that if a molecule has a change in dipole moment or has electron cloud polarizability, they are always IR and raman active? Are there change in dipole moment or cloud polarization that can't be detected by any spectrometer? – Jtl Oct 08 '19 at 00:18
  • 1
    @Jtl active mode doesn't implies the intensity of the abs. or scatt. to be high. The latter is always small and just lasers made raman possible. The effect is always there, just not detected using a standard light source. More close to your Q: for IR smaller is the el. dipole change weaker is the vibr. band intensity. For Raman the same though the change involves polarisability.In Raman, is even more complicated as coupling of vibrations with the el. transitions can lead to enhancement of bands. Vibrations within a chromophore can be greatly enhanced, so that remaining vibrations " vanishes". – Alchimista Oct 08 '19 at 09:34
  • for water/h2o which is both raman active and IR active. Is there a way to influence the molecules without it affecting either the dipole moment or polarization? How specificaly for water/h2o? – Jtl Oct 10 '19 at 01:51
  • 1
    "Is there a way to influence the molecules without it affecting either the dipole moment or polarization?".Not sure what you mean. If you mean if there is a normal mode that is silent, no. H2O has 3 normal modes which are both IR and Raman active. Its symmetry isn't high enough. – Alchimista Oct 10 '19 at 09:44
  • Using this glass vial on a 532nm raman spectrometer https://imagizer.imageshack.com/img924/4674/00uhyz.jpg How come at some distance I get this raman curve with some unknown peak at near 500 1/cm, https://imagizer.imageshack.com/img924/9114/MsOBdQ.png then at some distance, the peak vanishes https://imagizer.imageshack.com/img924/4361/eZs05H.png? What material do you think is the vial? Thanks. – Jtl Oct 10 '19 at 22:57
  • 1
    Tentative: That region might sense impurities that are hydrogen bound and vial walls might be involved. Assuming T to be reasonably constant you could see if the band change with time. Moreover I am not sure that the band really vanished if normalized to the main vibrational bands. You are at the detection boundary subtraction of noise and background might be tricky and you have basically two different setups (distance 1 and distance 2). – Alchimista Oct 11 '19 at 09:31
  • What do you mean by the "band change with time"? How do you adjust the time? I think by focusing the laser focal point right in the middle of the vial. It can sense just the water and not the wall and this is the second spectra with less disturbance from the wall? note the raman spectra of Boron silicate is exactly that region with the peak, see: https://www.researchgate.net/figure/Raman-spectra-of-borosilicate-glass-at-ambient-condition_fig1_234945622 – Jtl Oct 11 '19 at 11:06
  • 1
    I mean the time water stayed in the vial. If you are near a vial signal then it just matter of focusing / reflection and I would treat it as cut-off, meaning that region not very informative about water anymore. Note that 1) things are now totally unrelated to the Q 2) people can't guess details – Alchimista Oct 11 '19 at 11:21