I'm pretty confused about mass resolution. The resolution of my instrument is 1000 and I have two masses at 85.0639 and 85.0651. So, the difference is 0.0012. I know that (M1-M2)=M/R. But, I'm not sure if my spectrometer will be able to separate these ions.
-
2It is not very clear what those "1000" means. As resolution is usually meant the smallest difference distinguished by device reading. E.g. the classical, old school 2-arm mechanical scales had resolution 0.0001 g. – Poutnik Jan 03 '23 at 11:07
-
My bad, I meant mass resolution. – Jan 03 '23 at 11:20
-
2From your Q it seems 1/1000 is the relative resolution of the device, so for relative mass 85, resolution is 0.085 > 0.0012 so it seems you are out of luck. // If it is wrong, that it is probably your fault due bad description of scenario. – Poutnik Jan 03 '23 at 11:55
-
related/helpful: Resolution in Mass Spectrometry – uhoh Jan 04 '23 at 02:28
1 Answers
The mass resolution $R$ of a mass spectrometer is defined as
$$R = \frac{m}{\Delta{}m} $$
with $\Delta{}m$ the mass difference between two masses still be resolved, $\Delta{}m = m_1 - m_2$. Thus, if your spectrometer specs detail out $R = 1000$, then you record masses of $m_1 = 100.0$ and $m_2 = 100.1$ as two separate signals.*
In your example, $m_1 = 85.0639$ and $m_2 = 85.0651$, or a $|\Delta{}m| = 0.0012$. So you would need a resolution of
$$ R = \frac{85.0639}{0.0012} \approx 70887$$
to record these masses as separate signals. This number leans more toward a high-res spectrometer (e.g., to substitute a combustion analysis), than the simpler bench-tops routinely hyphenated at the exit of a HPLC.
* This omits how you define a peak in mass spectrometry. For simplification, this back on the envelope computation assumes zero-width of the signals and no partial overlap.
- 29,590
- 2
- 45
- 108
-
1+1, There is a non-ending argument about the terms spectroscopy vs. spectrometry :-) Your mass spectroscopy reminds me of my undergraduate analytical chemistry teacher held me outside the class and said that I was the only person in class who wrote mass spectroscopy in the exam. Why did you write that? It is mass spectrometry! I still see very established chemists arguing about it. – AChem Jan 03 '23 at 14:46
-
1@AChem The inconsistency (twice spectroscopy, twice spectrometry) is fixed. With "The term mass spectroscope continued to be used even though the direct illumination of a phosphor screen was replaced by indirect measurements with an oscilloscope." from wikipedia, I agree, since its recordings no longer rely on emitted/captured electromagnetic radiation, MS is better off as spectrometry. Google Ngram – Buttonwood Jan 03 '23 at 21:12